Laboratory errors, sample handling and quality control
Most wrong laboratory results start before the sample reaches the analyser; quality control catches the errors made inside the laboratory.
Key facts
- Errors are pre-analytical (before testing: wrong patient, timing, tube or handling), analytical (inside the instrument) or post-analytical (reporting and interpretation); the pre-analytical phase causes most of them.
- Accuracy is closeness to the true value (bias); precision is how closely repeated results agree (coefficient of variation = SD ÷ mean × 100).
- Percentage error = (measured − true value) ÷ true value × 100; each test has a total error the laboratory may allow.
- Internal quality control runs control samples of known value in every run and plots them on a Levey–Jennings chart; Westgard rules decide when to reject a run.
- External quality assessment (proficiency testing) compares the laboratory's results on the same samples with other laboratories.
Spurious results and their usual causes
| Result | Likely cause | Clue |
|---|---|---|
| High K⁺ (pseudohyperkalaemia) | Haemolysis, fist clenching, delayed or cold storage, very high platelets or white cells | Normal ECG; pink plasma; normal K⁺ in heparin plasma |
| High K⁺, low Ca²⁺, low ALP and Mg | EDTA contamination (wrong order of draw) | Impossible combination in a well patient |
| Low Na⁺ with normal osmolality | Very high lipids or proteins, indirect electrode | Lipaemic sample; normal Na⁺ on blood-gas analyser |
| Very high glucose, other values diluted | Sample drawn from the drip arm | Infusion running in the same arm |
| Low glucose | Unseparated sample without fluoride | Delay before analysis |
| Odd immunoassays (thyroid, troponin, hCG) | Biotin supplements, heterophile antibodies | Results do not fit the clinical picture |
In detail
- Tube types: EDTA for blood counts and HbA1c, citrate for coagulation, heparin for plasma chemistry, fluoride–oxalate for glucose (fluoride stops glycolysis by inhibiting enolase), plain or clot-activator tubes for serum.
- Glucose falls in an unseparated sample without fluoride because blood cells keep using it; potassium rises as cells leak.
- Haemolysis falsely raises potassium, LDH and AST; contamination with EDTA gives a falsely high potassium and low calcium.
- Standing upright, a long tourniquet and exercise raise proteins and protein-bound substances; lipaemia, jaundice and biotin supplements can interfere with some assays.
- A reference interval is usually the central 95 % of results from healthy people, so about 1 healthy person in 20 falls outside it on any one test.
- Also called
- Pre-analytical variables, Accuracy and precision
Clinical notes
- A very high potassium with no ECG changes in a well patient should first prompt a repeat sample to exclude haemolysis or EDTA contamination.
- Blood drawn from the arm with an intravenous infusion running is diluted by the infusion fluid.
- Running many tests on a healthy person will almost always produce some «abnormal» results by chance.
Common traps
- An abnormal result in a healthy person means disease: a reference range covers 95 % of healthy people, so with 20 tests about two in three healthy people have at least one result out of range.
- Accurate means precise: a method can repeat the same wrong value (precise but biased); accuracy is closeness to the true value.
- A very high potassium needs immediate treatment: in a well patient with a normal ECG first repeat the sample to exclude haemolysis or contamination, but act at once if the ECG changes.
- Quality control proves the patient's result is right: it watches the analytical leg; it cannot catch a wrong patient, a wrong tube or a delayed sample.
Test yourself
Think of the answer, then tap the question to see it.
What is the difference between accuracy and precision?
Accuracy is how close a result is to the true value; precision is how closely repeated measurements agree with each other.
A well patient has potassium 7.2 mmol/L with a normal ECG. What should be done first?
Repeat the sample, looking for haemolysis, EDTA contamination or delayed separation before treating.
A sample from a patient on an IV glucose infusion shows glucose 600 mg/dL, Na⁺ 128 and low values for everything else. What happened?
The sample was probably drawn from the drip arm and diluted by the infusion: glucose is high and the rest low. Redraw from the other arm.
A healthy 30-year-old has 20 routine tests; one ALT is just above range. Why is this not surprising?
Each range excludes 5 % of healthy people, so with 20 tests the chance of at least one out-of-range result is about 1 − 0.95²⁰ ≈ 64 %. Repeat it and judge the context.
Connections
Educational information, not medical advice. For any symptom, see a health professional.
Sources: How to Understand Your Lab Results: MedlinePlus Medical Test (MedlinePlus, US National Library of Medicine, 2026).
Spotted a mistake?We are always improving for the community. Tell us what is wrong and our team will review it.Report a mistake