Laboratory instruments: electrodes, analysers and dry chemistry
The machines behind common results: electrodes for pH, gases and electrolytes, automated analysers, dry-reagent strips and several kinds of electrophoresis.
Key facts
- A pH meter uses a glass electrode whose voltage changes with H⁺ concentration; it is calibrated with buffers of known pH. Blood-gas analysers add electrodes for pCO₂ and pO₂.
- Ion-selective electrodes measure Na⁺, K⁺, Cl⁻ and ionised Ca²⁺; they replaced flame photometry, which measured sodium and potassium by the light they give off in a flame.
- Direct electrodes read undiluted samples; indirect ones read a diluted sample and give a falsely low sodium when fats or proteins are very high (pseudohyponatraemia).
- Automated analysers handle many samples and tests at once (random access), with barcodes, fewer manual steps and fewer errors; most routine chemistry and immunoassays run on them.
- Dry chemistry puts the reagents in dried layers on a strip or slide; a drop of sample reacts and the colour is read by reflectance (urine dipsticks, glucose meters, point-of-care devices).
Serum protein electrophoresis: patterns to recognise
| Pattern | What the bands show | Think of |
|---|---|---|
| Narrow spike (M band) in γ or β | One clone making one immunoglobulin | Myeloma, MGUS, Waldenström; confirm with immunofixation |
| Broad γ rise (polyclonal) | Many clones responding | Chronic infection, autoimmune disease, cirrhosis |
| β–γ bridging | IgA rise filling the gap | Cirrhosis |
| Low albumin, high α₂ | Small proteins lost, α₂-macroglobulin kept | Nephrotic syndrome |
| Flat or absent α₁ | Little α₁-antitrypsin | α₁-antitrypsin deficiency |
| High α₁ and α₂ | Acute-phase proteins up | Acute inflammation |
In detail
- Atomic absorption and inductively coupled plasma mass spectrometry measure trace metals such as lead, copper and zinc.
- Electrophoresis variants: agarose for serum proteins, high-voltage electrophoresis for small molecules such as amino acids, capillary electrophoresis for serum proteins and haemoglobin variants, and immunofixation to type a monoclonal band.
- Point-of-care tests give fast results at the bedside but need the same quality control as laboratory tests.
- Tandem mass spectrometry measures many metabolites in one drop of blood and is the basis of expanded newborn screening.
- Also called
- pH meter and ion-selective electrodes, Autoanalysers and point-of-care testing
Clinical notes
- In severe hypertriglyceridaemia or myeloma, a low sodium from an indirect electrode may be false; a direct electrode (blood-gas analyser) or normal measured osmolality settles it.
- Glucose meters are less accurate at very low or very high values and with abnormal haematocrit; a laboratory sample confirms extreme readings.
Common traps
- Every analyser reads sodium the same way: indirect electrodes read a diluted sample and give a falsely low sodium with very high lipids or proteins; direct electrodes (blood-gas analysers) do not.
- The blood-gas bicarbonate is measured: it is calculated from pH and pCO₂; the chemistry panel measures total CO₂.
- Electrophoresis types the M band: it only shows a band; immunofixation names its heavy and light chains.
- A glucose meter is as accurate as the laboratory: strips lose accuracy at very low or high values and with abnormal haematocrit; extreme readings are confirmed in the laboratory.
Test yourself
Think of the answer, then tap the question to see it.
Why can a laboratory report a low sodium in a patient with very high triglycerides when the true sodium is normal?
Indirect electrodes dilute the sample; the lipid takes up volume, so the calculated sodium per litre of plasma is falsely low (pseudohyponatraemia).
What does dry chemistry mean?
Reagents are dried in layers on a strip or slide; the sample dissolves them and the colour change is read, as on urine dipsticks and glucose meters.
A 70-year-old with anaemia and back pain has a narrow spike in the γ region on serum protein electrophoresis. What is the next laboratory test?
Immunofixation, to identify the monoclonal immunoglobulin, with serum free light chains; the picture suggests multiple myeloma.
On the same blood, the gas report gives HCO₃⁻ 18 mmol/L and the chemistry panel total CO₂ 20 mmol/L. Is one of them wrong?
Not necessarily: the gas HCO₃⁻ is calculated from pH and pCO₂, while total CO₂ is measured and includes dissolved CO₂, so it runs 1–2 mmol/L higher.